Teledyne T703 VCR User Manual


 
Teledyne API T703/T703U Calibrator Operation Manual Calibration and Verification
133
6.2.2. VIEWING O
3
GENERATOR CALIBRATION POINTS
To view these calibration points, press:
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
Make sure that the
calibrator is in standby
mode.
SETUP X.X ENTER PASSWORD
818 ENTREXIT
Toggle to enter the correct
PASSWORD
DIAG SIGNAL I/O
PREV NEXT ENTR EXIT
DIAG O3 GEN CALIBRATION
PREV NEXT ENTR EXIT
Continue pressing NEXT until ...
DIAG O3 GEN CALIBRATION
CAL PNTS EXIT
Toggle to move between
calibration points
DIAG O3GEN 1) 500 MV, 5.0 MIN, ON
PREV NEXT INS DEL EDIT PRNT EXIT
STANDBY ACT =STANDBY
<TST TST> GEN STBY SEQ SETUP
SETUP X.X PRIMARY SETUP MENU
O3 SEQ CFG CLK PASS MORE EXIT
07223C DCN6572