Teledyne T703 VCR User Manual


 
Getting Started Teledyne API T703/T703U Calibrator Operation Manual
36
3.2.5. CONNECTING THE CONTROL OUTPUTS
The calibrator is equipped with 12 opto-isolated, digital control outputs. These outputs
are activated by the user-programmable, calibration sequences (see Section 4.5.1.6 for
instructions on assigning the control outputs to specific calibration sequences)
These outputs may
be used to interface with devices that accept logic-level digital
inputs, such as programmable logic controllers (PLCs), dataloggers, or digital
relays/valve drivers.
They are accessed via a 14-pin connector on the calibrator’s rear panel (see Figure 3-4).
CONTROL OUTPUTS
1 2 3 4 5 6 7 8 9 10 11 12 E
Figure 3-12: Digital Control Output Connector
NOTE
Most PLCs have internal provisions for limiting the current the input will draw. When connecting to a
unit that does not have this feature, external resistors must be used to limit the current through the
individual transistor outputs to 50mA (120 for 5V supply).
The pin assignments for the control outputs are:
Table 3-5: Control Output Pin Assignments
PIN # STATUS DEFINITION CONDITION
1 - 12 Outputs 1 through 12 respectively Closed if the sequence or sequence step activating output is operating
E Emitter BUSS The emitters of the transistors on pins 1 to 8 are bussed together.
Digital Ground The ground level from the calibrator’s internal DC power supplies.
07223C DCN6572