Teledyne API T703/T703U Calibrator Operation Manual Operating the Calibrator
59
4.5.1.5. Setting Up Control Inputs for a Sequence
The calibrator’s control inputs allow the entire sequence to be triggered from an external
source. This feature allows the calibrator to operate in a slave mode so that external
control sources, such as a data logger can initiate the calibration sequences.
Each of the calibrator’s control outputs, located on the back of the instrument (see
Figure 3-2)
12 separate ON/OFF switches assigned to separate calibration sequences or;
A 12-bit wide bus allowing the user to define activation codes for up to 4095
separate calibration sequences.
To assign a CC INPUT pattern / code to a particular sequence, press.