4.6.7. PRESSURE CALIBRATION
A sensor at the exit of the sample chamber continuously measures the pressure of the
sample gas. This data is used to compensate the final H
2
S concentration calculation for
changes in atmospheric pressure when the instrument’s TPC feature (see Section 10.7.3)
is turned on and is stored in the CPU’s me
mory as the test function PRES (also viewable
via the front panel).
To cause the analyzer to measure and record a value for PRES, press:
Exit at
any time
to return
to main
the
SETUP
menu
DIAG SIGNAL I / O
NEXT ENTR EXIT
Repeat Pressing NEXT until . . .
DIAG PRESSURE CALIBRATION
PREV NEXT ENTR EXIT
DIAG PCAL
A
CTUAL PRES :27.20 IN-HG-
A
2 7 .2 0 ENTR EXIT
ENTR accepts the
new value
EXIT ignores the new
value
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL SETUP
SAMPLE ENTER SETUP PASS : 818
8 1 8 ENTR EXIT
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
Adjust these values until the
displayed pressure equals the
pressure measured by the
independent pressure meter.
07266B DCN6485