Teledyne API T/E-Series Models 101, 102, and 108 Software Documentation (PN05492D DCN6485)
A-12
Setup Variable
Numeric
Units
Default
Value
Value Range Description
CONC_PRECISION — 1 AUTO,
0,
1,
2,
3,
4
Number of digits to display
to the right of the decimal
point for concentrations on
the display. Enclose value in
double quotes (“) when
setting from the RS-232
interface.
CLOCK_ADJ Sec./Day 0 -60–60 Time-of-day clock speed
adjustment.
SERVICE_CLEAR — OFF OFF
ON
ON resets the service
interval timer.
TIME_SINCE_SVC Hours 0 0–500000 Time since last service.
SVC_INTERVAL Hours 0 0–100000 Sets the interval between
service reminders.
Medium Access Level Setup Variables (929 password)
LANGUAGE_SELECT — ENGL ENGL,
SECD,
EXTN
Selects the language to use
for the user interface.
Enclose value in double
quotes (“) when setting from
the RS-232 interface.
LATCH_WARNINGS — ON ON, OFF ON enables latching warning
messages; OFF disables
latching
DAYLIGHTSAVING_ENABLE — ON ON, OFF ON enables Daylight Saving
Time (DST) change; OFF
disables DST.
MAINT_TIMEOUT Hours 2 0.1–100 Time until automatically
switching out of software-
controlled maintenance
mode.
MEASURE_PERIOD Minutes 10 1–60 Length of time to measure
each gas.
MEASURE_DELAY Minutes 3 0.1–20 How long to defer sampling
after switching streams.
FLUSH_DURATION
7
Seconds 10 0.1–100 Length of time to flush prior
to measuring each gas.
CONV_TIME — 33 MS 33 MS,
66 MS,
133 MS,
266 MS,
533 MS,
1 SEC,
2 SEC
Conversion time for PMT and
UV detector channels.
Enclose value in double
quotes (“) when setting from
the RS-232 interface.
DWELL_TIME Seconds 1 0.1–10 Dwell time before taking
each sample.
FILT_SIZE Samples 240 1–480 Moving average filter size.
FILT_ASIZE Samples 20 1–100 Moving average filter size in
adaptive mode.
07266B DCN6485