will automatically be assigned an IP address by the DHCP server (Section 4.10.6.2). This
configuration is useful for quickly getting an instrument up and running on a network.
However, for permanent Ethernet connections, a static IP address should be used. Section
4.7.6.2 below details how to configure the instrume
nt with a static IP address.
NOTE
It is a good idea to check these settings the first time you power up your analyzer after it has been
physically connected to the LAN/Internet to make sure that the DHCP has successfully downloaded the
appropriate information from you network server(s).
If the gateway IP, instrument IP and the subnet mask are all zeroes (e.g. “0.0.0.0”), the DCHP was not
successful.
You may have to manually configure the analyzer’s Ethernet properties.
See your network administrator.
To view the LAN/Internet default configuration properties, press:
SAMPLE EN
T
ER SETUP PASS : 818
8 1 8 ENTR EXIT
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL SETUP
SETUP X.X COMMUNICATIONS MENU
ID INET COM1 COM2 EXIT
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
SETUP X.X DHCP: ON
SET> EDIT EXIT
SETUP X.X INST IP: 0.0.0.0
<SET SET> EXIT
SETUP X.X GATEWAY IP: 0.0.0.0
<SET SET> EXIT
SETUP X.X SUBNET MASK: 0.0.0.0
<SET SET> EXIT
SETUP X.X TCP PORT: 3000
<SET SET> EDIT EXIT
SETUP X.X HOSTNAME: T101
<SET EDIT EXIT
EDITING
disabled
when
DHCP is
ON
From this point on,
EXIT returns to
COMMUNICATIONS
MENU
Do not alter
unless
directed to by
Teledyne
AP I ’ s
Customer
Service
personnel