Principles Of Operation Model T101 Instruction Manual
274
10.7. SOFTWARE OPERATION
The T101 H
2
S analyzer is at its heart a high performance, 386-based microcomputer
running MS-DOS. Inside the DOS shell, special software developed by Teledyne API
interprets user commands via the various interfaces, performs procedures and tasks,
stores data in the CPU’s various memory devices and calculates the concentration of the
sample gas.
Windows CE
API FIRMWARE
Analyzer Operations
Calibration Procedures
Configuration Procedures
Autonomic Systems
Diagnostic Routines
Memory Handling
DAS Records
Calibration Data
System Status Data
Interface Handling
Sensor input Data
Display Messages
Touchscreen
Analog Output Data
RS232 & RS485
External Digital I/O
Measurement
Algorithm
(H
2
S & SO
2
)
ANALYZER
HARDWARE
PC/104 BUS
PC/104 BUS
Figure 10-20. Basic Software Operation
10.7.1. ADAPTIVE FILTER
The T101 H
2
S analyzer software processes sample gas measurement and reference data
through a built-in adaptive filter built into the software. Unlike other analyzers that
average the sensor output signal over a fixed time period, the T101 calculates averages
over a set number of samples. During operation, the software automatically switches
between two filters of different lengths based on the conditions at hand.
During conditions of constant or nearly constant concentration the software computes an
average of the last 240 samples. This provides the calculation portion of the software
with smooth stable readings. If a rapid change in concentration is detected, the adaptive
filter switches modes and only averages the last 48 samples. This allows the analyzer to
respond to the rapidly changing concentration more quickly. Once triggered, the short
filter remains engaged for a fixed time period to prevent chattering.
Two conditions must be simultaneously met to switch to the short filter. First the
instantaneous concentration must exceed the average in the long filter by a fixed
amount. Second, the instantaneous concentration must exceed the average in the long
filter by a portion, or percentage, of the average in the long filter.
If necessary, the filter lengths of these two modes may be changed to any value between
1 and 1000 samples. Long sample lengths provide better signal to noise rejection, but
poor response times. Conversely shorter filter lengths result in poor signal to noise
rejection, but quicker response times.
07266B DCN6485