Teledyne T101 TV Converter Box User Manual


 
Teledyne API – T101 Operation Manual Calibration Procedures
173
6.5. MANUAL CALIBRATION WITH IZS OPTION
The following section describes the basic method for manually calibrating the Model
T101 analyzer in H
2
S measurement mode using the IZS option. The same method may
be used to calibrate the T101 analyzers configured for SO
2
measurement but requires that
an SO
2
permeation tube be installed in the instrument instead of the standard H
2
S tube.
(SO
2
permeation tubes differ from H
2
S tubes and can be purchased directly from a
manufacturer).
See Section 6.8 for instructions on calibrating analyzers configured for multigas
measurement m
ode.
Under the best conditions, the accurac
y off the H
2
S effusion rate of the IZS option’s
permeation tube is about ±5%. This can be subject to significant amounts of drift as the
tube ages and the amount of H
2
S contained in the tube is depleted. Whereas this may be
sufficient for calibrating instrument configured for H
2
S measurement and for informal
calibration checks of instruments measuring SO
2
, it is not adequate for formal SO
2
calibrations and is not approved for use by the US EPA as a calibration source for
calibrating SO
2
monitoring equipment.
For applications where more stringent calibration requirements are specified for an
instrument with an IZS option installed the following provisions must be followed.
1. Zero air and span gas must be supplied to the analyzer through the
sample gas inlet as depicted in Figure 6-1.
2. Th
e calibration procedure must be initiated using the CAL button (not the
CALZ and CALS buttons) using the procedure defined in Section 6.2.
3. Using the CA
L button does not activate the zero/span or sample/cal
valves of the IZS option, thus allowing the introduction of zero air and
sample gas through the sample port from more accurate, external sources
such as a calibrated bottle of H
2
S and SO
2
or a Model T700 Dilution
Calibrator.
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL CALZ CALS SETUP
Use for formal
calibration
operations.
Use only for
informal calibration
checks.
07266B DCN6485