Operating Instructions Teledyne API – T101 Operation Manual
80
4.5.1. SETTING THE GAS MEASUREMENT MODE
If the switching valves software is activated, the T101 can be set to one of three gas
measurement modes:
H
2
S
The sample gas stream is stripped of any ambient SO
2
by a special
chemical scrubber, then passed through a catalytic converter that changes
the H
2
S present into SO
2
which is then measured using the UV
Fluorescence method
SO
2
The sample gas stream bypasses the SO
2
Scrubber and catalytic converter
allowing the only ambient SO
2
to be measured.
H
2
S – SO
2
The switching valve alternates the gas stream between the two paths at
regular intervals allowing the instrument to measure both gases.
To select one of these three measurement modes (see Section 10.3.2 for additional
details), press:
SETUP X.X 0 ) MEASURE_MODE=H2S
NEXT JUMP EDIT PRNT EXIT
EXIT ignores the new
setting.
ENTR accepts the
new setting.
SAMPLE ENTER SETUP PASS : 818
8 1 8 ENTR EXIT
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SETUP X.X
SECONDARY SETUP MENU
COMM VARS DIAG EXIT
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL SETUP
SETUP X.X MEASURE MODE: H2S
PREV ENTR EXIT
SETUP X.X MEASURE MODE: H2S-SO2
PREV NEXT ENTR EXIT
SETUP X.X MEASURE MODE: SD2
NEXT ENTR EXIT
H
2
S mode is the
default mode.
Press the PREV
and NEXT buttons
to scroll among gas
measurement
mode choices.
07266B DCN6485