Teledyne API T/E-Series Models 101, 102, and 108 Software Documentation (PN05492D DCN6485)
A-11
APPENDIX A-2: Setup Variables For Serial I/O, S/W Version C.7 (E-Series), 1.0.5 (T-
Series)
Table A-1: Setup Variables
Setup Variable
Numeric
Units
Default
Value
Value Range Description
Low Access Level Setup Variables (818 password)
MEASURE_MODE — SO2-H2S,
TRS
6
SO2,
H2S,
TRS
6
,
SO2-H2S,
SO2-TRS
6
,
H2S-TRS
6
,
SO2-H2S-
TRS
6
,
SO2-H2S-
REMOTE,
H2S-SO2-
REMOTE
Gas measurement mode.
Enclose value in double
quotes (") when setting from
the RS-232 interface.
CAL_GAS — DEF DEF,
SO2,
H2S
Selects calibration gas (i.e.
valve position, as opposed to
slope/offset). DEF selects
default behavior, in which
valve position and
slope/offset are the same.
Enclose value in double
quotes (") when setting from
the RS-232 interface.
DAS_HOLD_OFF Minutes 15 0.5–20 Duration of DAS hold off
period.
TPC_ENABLE — ON OFF, ON ON enables temperature and
pressure compensation; OFF
disables it.
50 RCELL_SET ºC
Warnings:
45–55
30–70 Reaction cell temperature
set point and warning limits.
50 IZS_SET ºC
Warnings:
45–55
30–70 IZS temperature set point
and warning limits.
DYN_ZERO — OFF OFF, ON ON enables contact closure
dynamic zero; OFF disables
it.
DYN_SPAN — OFF OFF, ON ON enables contact closure
dynamic span; OFF disables
it.