Teledyne API – T101 Operation Manual Calibration Procedures
183
It is recommended that, if dynamic calibration is enabled, the analyzer’s test functions,
slope and offset values be checked frequently to assure high quality and accurate data
from the instrument.
6.9.1. AUTOCAL OF INSTRUMENTS IN INDEPENDENT OR AUTO
REPORTING RANGE MODES
If the analyzer is being operated in IND or AUTO Range mode, then the High and Low
ranges must be specified as part of the Auto Cal set up. This parameter appears at the end
of the programming sequences after the CALIBRATE: ON/OFF parameter is set. For
example:
SETUP X.X RANGE TO CAL: LO
W
<SET EDIT EXIT
EXIT returns
to the SETUP
Menu
SETUP X.X CALIBRATE: ON
<SET SET> EDIT EXIT
SETUP X.X SEQ 2) ZERO–SPAN, 2:00:30
PREV NEXT MODE SET EXIT
Sequence Delta Time
MODE Delta Days
Follow standard AutoCal programming process to this
point, then …
SETUP X.X RANGE TO CAL: LOW
LOW HIGH ENTR EXIT
SETUP X.X RANGE TO CAL: HIGH
LOW HIGH ENTR EXIT
SETUP X.X RANGE TO CAL: HIGH
<SET EDIT EXIT