Calibration Procedures Teledyne API – T101 Operation Manual
182
Toggle to
set
number of
days
between
procedures
(1-367)
Toggle to set
delay time for
each iteration
of the
sequence:
HH:MM
(0 – 24:00)
SETUP X.X DELTA DAYS: 1
<SET SET> EDIT EXIT
SETUP X.X DELTA DAYS: 1
0 0 2 ENTR EXIT
SETUP X.X DELTA DAYS:2
<SET SET> EDIT EXIT
SETUP X.X DELTA TIME00:00
<SET SET> EDIT EXIT
SETUP X.X DELTA TIME: 00:00
0 0 :3 0 ENTR EXIT
SETUP X.X DELTA TIEM:00:30
<SET SET> EDIT EXIT
SETUP X.X STARTING TIME:14:15
<SET SET> EDIT EXIT
EXIT returns
to the SETUP
Menu
SETUP X.X DURATION:15.0 MINUTES
<SET SET> EDIT EXIT
SETUP X.X DURATION 15.0MINUTES
3 0 .0 ENTR EXIT
SETUP X.X DURATION:30.0 MINUTES
<SET SET> EDIT EXIT
Toggle to
set
duration for
each
iteration of
the
sequence:
Set in
Decimal
minutes
from
0.1 – 60.0
SETUP X.X C
A
LIBRATE: OFF
<SET SET> EDIT EXIT
SETUP X.X CALIBRATE: OFF
ON ENTR EXIT
SETUP X.X CALIBRATE: ON
<SET SET> EDIT EXIT
Toggle
button
between
Off and
ON
SETUP X.X SEQ 2) ZERO
SPAN, 2:00:3
0
PREV NEXT MODE SET EXIT
SAMPLE ENTER SETUP PASS : 818
8 1 8 ENTR EXIT
SETUP X.X PRIMARY SETUP MENU
CFG ACAL DAS RNGE PASS CLK MORE EXIT
SETUP X.X SEQ 1) DISABLED
NEXT MODE EXIT
SETUP X.X SEQ 2) DISABLED
PREV NEXT MODE EXIT
SETUP X.X MODE: DISABLED
NEXT ENTR EXIT
SETUP X.X MODE: ZERO
PREV NEXT ENTR EXIT
SETUP X.X MODE: ZERO
SPAN
PREV NEXT ENTR EXIT
SETUP X.X SEQ 2) ZERO–SPAN, 1:00:00
PREV NEXT MODE SET EXIT
SETUP X.X TIMER ENABLE: ON
SET> EDIT EXIT
SETUP X.X STARTING DATE: 01–JAN–02
<SET SET> EDIT EXIT
SETUP X.X STARTING DATE: 01–JAN–02
0 4 SEP 0 3 ENTR EXIT
Toggle to set
day, month &
year:
Format :
DD-MON-Y
Y
SETUP X.X STARTING DATE: 04–SEP–03
<SET SET> EDIT EXIT
Default
value is
ON
SETUP X.X STARTING DATE: 04–SEP–03
<SET SET> EDIT EXIT
SETUP X.X STARTING TIME:00:00
<
S
ET
S
ET> EDIT EXIT
Toggle to set
time:
Format : HH:MM
Thisisa24
hr
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL CALZ CALS SETUP
With dynamic calibration turned on, the state of the internal setup variables dyn_Span and
DYN_ZERO is set to ON and the instrument will reset the slope and offset values for the
H
2
S response each time the AutoCal program runs. This continuous re-adjustment of
calibration parameters can often mask subtle fault conditions in the analyzer.
07266B DCN6485