Teledyne T101 TV Converter Box User Manual


 
Calibration Procedures Teledyne API – T101 Operation Manual
184
6.9.2. AUTOCAL OF INSTRUMENTS IN MULTIGAS
MEASUREMENT MODE
If the analyzer is being operated in multigas measurement mode, auto cal can only be
programmed to calibrate the instrument with one of the available gases.
NOTE
Make sure that the gas for which the AutoCal sequence has been programmed is being
supplied to the analyzer at the time the autocal program initiates.
If the wrong gas is being supplied to the instrument, the analyzer will be mis-calibrated.
If the instrument is in multigas measurement mode and the SNGL reporting range mode
is selected, the AutoCal sequence is programmed as described above. If the reporting
range of the T101 must is set for either IND or AUTO. the gas type and range to be
measured must be specified. This parameter appears at the end of the programming
sequences after the CALIBRATE: ON/OFF parameter is set. For example:
SETUP X.X GAS-RANGE: H2S-HIGH
<SET EDIT EXIT
SETUP X.X CALIBRATE: ON
<SET SET> EDIT EXIT
Follow standard AutoCal programming process to this
point, then …
SETUP X.X GAS TO CAL: H2S
SO2 H2S ENTR EXIT
SETUP X.X Gas TO CAL: SO2
LOW HIGH ENTR EXIT
EXIT returns
to the SETUP
Menu
SETUP X.X SEQ 2) ZERO
SPAN, 2:00:30
PREV NEXT MODE SET EXIT
Sequence Delta Time
MODE Delta Days
SETUP X.X RANGE TO CAL: LOW
LOW HIGH ENTR EXIT
SETUP X.X RANGE TO CAL: HIGH
LOW HIGH ENTR EXIT
SETUP X.X GAS-RANGE: SO2-LOW
<SET EDIT EXIT
07266B DCN6485