Teledyne T101 TV Converter Box User Manual


 
Getting Started Teledyne API – T101 Operation Manual
42
PMT
EXHAUST THROUGH OUTER
LAYER OF KICKER
REACTION CELL PURGE
FL O W
SENSOR
FLOW / PRESSURE
SENSOR PCA
SAMPLE
FILTER
INSTRUMENT CHASSIS
EXH AUST GAS
OUTLET
VACUUM MANIFOLD
SAMPLE GAS
INLET
KIC KER EX HAUST TO PU MP
PUMP
FL O W
CO NTRO L
ASSY
SAMPLE
PRESSURE
SENSOR
UV
LAM P
HYDROCARBON
SCRUBBER
(KICKE R)
SPAN GAS INLET
ZERO AIR INLET
SAMPLE
CHAMBER
CO M
NC NO
SO
2
Scrubber
MOLYBDENUM
CONVERTER
SO
2
H
2
S
H2S / SO2
MODE VALVE
Gas Flow when m ultigas version of
Analyzer is measuring SO
2.
Figure 3-13. Pneumatic Diagram of the T101 Standard Configuration
Table 3-8. H
2
S – SO
2
Switching Valve Operating States
GAS
MODE
CONDITION OF H
2
S –SO
2
SWITCHING
VALVE
VALVE PORT
CONNECTION
H
2
S
Open to SO
2
Scrubber and Molybdenum
Converter
COM NO
SO
2
Open to directly to Sample Chamber. Bypasses
SO
2
Scrubber and Molybdenum Converter
COM NC
H
2
S –SO
2
Switches between above two states every 10
minutes.
- -
1. Attach the 1/4" exhaust line to the EXHAUST port of the analyzer and to
the inlet port of the pump.
07266B DCN6485