Teledyne T101 TV Converter Box User Manual


 
Teledyne API T/E-Series Models 101, 102, and 108 Software Documentation (PN05492D DCN6485)
A-23
Table A-3: Test Measurements
TEST Measurement Message Text DESCRIPTION
RANGE RANGE=500.0 PPB
3
D/A range in single or auto-range
modes.
RANGE1 RANGE1=500.0 PPB
3
D/A #1 range in independent range
mode.
RANGE2 RANGE2=500.0 PPB
3
D/A #2 range in independent range
mode.
RANGE3
5
RANGE3=500.0 PPB
3
D/A #3 range in independent range
mode.
STABILITY SO2 STB=11.4 PPB
3
Concentration stability #1.
STABILITY2
8
SO2 STB2=6.3 PPB
3
Concentration stability #2.
RESPONSE
2
RSP=1.11(0.00) SEC Instrument response. Length of
each signal processing loop. Time in
parenthesis is standard deviation.
OXYFLOW
6
OXY FLOW=150 CC/M Oxygenator flow rate
SAMPFLOW SAMP FL=700 CC/M Sample flow rate.
SAMPPRESS PRES=29.9 IN-HG-A Sample pressure.
PMTDET PMT=762.5 MV Raw PMT reading.
NORMPMTDET NORM PMT=742.9 MV PMT reading normalized for
temperature, pressure, auto-zero
offset, but not range.
UVDET UV LAMP=3457.6 MV UV lamp reading.
STABILITYUV
9
UV STB=5.607 MV UV lamp stability reading.
LAMPRATIO LAMP RATIO=100.0 % UV lamp ratio of current reading
divided by calibrated reading.
STRAYLIGHT STR. LGT=0.1 PPB Stray light offset.
DARKPMT DRK PMT=19.6 MV PMT dark offset.
DARKLAMP DRK LMP=42.4 MV UV lamp dark offset.
SO2SLOPE SO2 SLOPE=1.000 Slope for current range, computed
during zero/span calibration.
SO2OFFSET SO2 OFFS=0.0 MV Offset for current range, computed
during zero/span calibration.
H2SSLOPE H2S SLOPE=1.000 Slope for current range, computed
during zero/span calibration.
H2SOFFSET H2S OFFS=0.0 MV Offset for current range, computed
during zero/span calibration.
TRSSLOPE
5
TRS SLOPE=1.000 Slope for current range, computed
during zero/span calibration.
TRSOFFSET
5
TRS OFFS=0.0 MV Offset for current range, computed
during zero/span calibration.
HVPS HVPS=650 VOLTS High voltage power supply output.
RCELLDUTY
2
RCELL ON=0.00 SEC Reaction cell temperature control
duty cycle.
RCELLTEMP RCELL TEMP=52.1 C Reaction cell temperature.
BOXTEMP BOX TEMP=35.5 C Internal chassis temperature.
07266B DCN6485