Teledyne T101 TV Converter Box User Manual


 
Teledyne API – T101 Operation Manual Operating Instructions
69
4.4.4.4. Single Range Mode (SNGL)
The default range mode for the analyzer is single range, in which all analog concentration
outputs are set to the same reporting range. This reporting range can be set to any value
between 5.0 ppb and 20 000 ppb.
While the two outputs always have the same reporting range, the span and scaling of their
electronic signals may also be configured for different differently (e.g., A1 = 0-10 V; A2
= 0-0.1 V).
To select SNGLE range mode and to set the upper limit of the range, press:
SAMPLE ENTER SETUP PASS : 818
8 1 8 ENTR EXIT
SETUP C.3
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SETUP C.3 RANGE CONTROL MENU
MODE
SET UNIT EXIT
SETUP C.3 RANGE MODE: SNGL
SNGL
IND AUTO ENTR EXIT
EXIT x 2 returns
to the main
SAMPLE display
SETUP C.3 RANGE MODE: SNGL
SNGL
IND AUTO ENTR EXIT
SETUP C.3 RANGE CONTROL MENU
MODE
SET UNIT EXIT
SETUP C.3 RANGE: 500.0 Conc
0 0 5 0 0 .0 ENTR EXIT
SETUP C.3 RANGE CONTROL MENU
MODE SET UNIT EXIT
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL SETUP
NOTE
On analyzers with the multigas option activated (see Sections 4.5.1 and 6.8) the concentration value will
switch back and forth between from “H2S=XXX.X” to “SO2=XXX.X” depending on which gas is currently
being measured.