Teledyne API T/E-Series Models 101, 102, and 108 Software Documentation (PN05492D DCN6485)
A-15
Setup Variable
Numeric
Units
Default
Value
Value Range Description
PHYS_RANGE1 PPM 2 0.1–2500 Low pre-amp range.
PHYS_RANGE2 PPM 20 0.1–2500 High pre-amp range.
CONC_RANGE1 Conc 500 0.1–50000 D/A concentration range 1.
CONC_RANGE2 Conc 500 0.1–50000 D/A concentration range 2.
CONC_RANGE3
6
Conc 500 0.1–50000 D/A concentration range 3.
150 OXY_FLOW_SET
4
cc/m
Warnings:
50–300
0–500 Nominal oxygenator flow set
point and warning limits.
OXY_FLOW_SLOPE
4
1 0.5–1.5 Slope term to correct
oxygenator flow rate.
700 SAMP_FLOW_SET cc/m
Warnings:
350–1200
0–1200 Sample flow set point for
flow calculation and warning
limits.
SAMP_FLOW_SLOPE 1 0.5–1.5 Sample flow slope correction
factor (adjusted flow =
measured flow x slope).
SFLOW_FILT_SIZE Samples 50 1–200 Sample flow adaptive
moving average filter size.
29.92 SAMP_PRESS_SET "Hg
Warnings:
15–35
0–100 Sample pressure set point
for pressure compensation
and warning limits.
CONV_TYPE
2
MOLY NONE, MOLY Converter type.
315 CONV_SET
2
ºC
Warnings:
310–320
0–350 Converter temperature set
point and warning limits.
30 BOX_SET ºC
Warnings:
8–50
5–60 Box temperature warning
limits. Set point is not used.
7,
15
9
PMT_SET ºC
Warnings:
2–12,
2–20
9
0–40 PMT temperature set point
and warning limits.
07266B DCN6485