Teledyne T101 TV Converter Box User Manual


 
Teledyne API – T101 Operation Manual Operating Instructions
89
To automatically calibrate an single analog channel, activate the ANALOG I/O
CONFIGURATION MENU (see Section 4.6.3), then press:
DIAG AIO CONC_OUT_2 CALIBR
A
TED: NO
<SET CAL EXIT
DIAG AIO
A
UTO CALIBRATING CONC_OUT_2
DIAG AIO CONC_OUT_2 CALIBRATED: YES
<SET CAL EXIT
DIAG ANALOG I / O CONFIGURATION
PREV NEXT ENTR EXIT
DIAG AIO AOUTS CALIBRATED: NO
< SET> CAL EXIT
DIAG AIO CONC_OUT_2 RANGE: 5V
SET> EDIT EXIT
DIAG AIO CONC_OUT_2:5V, CAL
< SET SET> EDIT EXIT
DIAG AIO CONC_OUT_2 REC OF
S
: 0 mV
< SET SET> EDIT EXIT
DIAG AIO CONC_OUT_2 AUTO CAL: ON
< SET SET> EDIT EXIT
EXIT to Return
to the main
Sample Display
Press SET
>
to select the
Analog Output channel to
be configured. Then Press
EDIT to continue
07266B DCN6485