4.7.6.2. Manually Configuring the Ethernet with Static IP Addresses
1. Connect a cable from the analyzer’s Ethernet port to a Local Area Network
(LAN) or Internet port.
2. Access Ethernet configuration through: SETUP>MORE>COMM>INET.
3. Follow the setup sequence as shown in the illustrations that follow, and
edit the Instrument and Gateway IP addresses and Subnet Mask to the
desired settings. (From the computer, enter the same information through
an application such as HyperTerminal).
Table 4-17. LAN/Internet Default Configuration Properties
PROPERTY DEFAULT STATE DESCRIPTION
DHCP
ON
This displays whether the DHCP is turned ON or
OFF. Press EDIT and toggle ON for automatic
configuration after first consulting network
administrator.
INSTRUMENT IP
ADDRESS
0.0.0.0
This string of four packets of 1 to 3 numbers each
(e.g. 192.168.76.55.) is the address of the
analyzer itself.
GATEWAY IP
ADDRESS
0.0.0.0
Can only be edited when DHCP is set to OFF.
A string of numbers very similar to the Instrument
IP address (e.g. 192.168.76.1.) that is the address
of the computer used by your LAN to access the
Internet.
SUBNET MASK
0.0.0.0
Can only be edited when DHCP is set to OFF.
Also a string of four packets of 1 to 3 numbers each
(e.g. 255.255.252.0) that identifies the LAN to
which the device is connected.
All addressable devices and computers on a LAN
must have the same subnet mask. Any
transmissions sent to devices with different subnets
are assumed to be outside of the LAN and are
routed through the gateway computer onto the
Internet.
TCP PORT
1
3000
This number defines the terminal control port by
which the instrument is addressed by terminal
emulation software, such as Internet or Teledyne
API’s APICOM.
HOST NAME
T101
The name by which your analyzer will appear when
addressed from other computers on the LAN or via
the Internet. To change, see Section 4.7.6.3.
1
Do not change the setting for this property unless instructed to by Teledyne API’s Technical
Support personnel.