Getting Started Teledyne API – T101 Operation Manual
54
STEP TWO: Set the expected H
2
S span gas concentration.
This sequence causes the
analyzer to prompt for the
expected H
2
S span
concentration.
M-P CAL H2S SPAN CONC: 450.0 Con
c
0 0 0 4 5 0 .0 ENTR EXIT
M-P CAL RANGE = 500.000 PPB H2S =X.XXX
< TST TST > ZERO CONC EXIT
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL SETUP
The H
2
S span
concentration values
automatically default to
450.0 Conc.
To change this value to
the actual concentration of
the span gas, enter the
number by pressing the
button under each digit
until the expected value
appears.
EXIT ignores the new setting
and returns to the previous
display.
ENTR accepts the new setting
and returns to the
previous display..