Teledyne API – T101 Operation Manual Operating Instructions
135
To configure the parameters for a specific data parameter, press:
FROM THE EDIT DATA PARAMETER MENU
(see previous section)
SETUP X.X 0) PARAM=S2SCN1, MODE=AVG
PREV NEXT INS DEL EDIT EXIT
SETUP X.X PARAMETERS:S2SCN1
SET> EDIT EXIT
SETUP X.X SAMPLE MODE:AVG
<SET SET> EDIT EXIT
SETUP X.X PARAMETERS: S2SCN1
PREV NEXT ENTR EXIT
Cycle through list of available
Parameters.
ENT
R
accepts the new
setting and returns to the
previous menu.
EXIT ignores the new setting
and returns to the previous
SETUP X.X SAMPLE MODE: INST
INST AVG MIN MAX EXIT
Press the key for the desired mode
SETUP X.X PRECISION: 1
<SET SET> EDIT EXIT
SETUP X.X PRECISION: 1
1 EXIT
Set for 0-4
SETUP X.X STORE NUM. SAMPLES: OFF
<SET EDIT EXIT
SETUP X.X STORE NUM. SAMPLES: OFF
OFF ENTR EXIT
Turn ON or OFF
<SET Returns to
previous
Functions
4.8.2.5. Sample Period and Report Period
The DAS defines two principal time periods by which sample readings are taken and
permanently recorded:
SAMPLE PERIOD: Determines how often DAS temporarily records a sample reading
of the parameter in volatile memory. The SAMPLE PERIOD is set to one minute by
default and generally cannot be accessed from the standard DAS front panel menu, but is
available via the instrument’s communication ports by using APICOM or the analyzer’s
standard serial data protocol.
07266B DCN6485