4.4.4.5. Independent Range Mode (IND)
Selecting independent range mode allows the A1 and A2 outputs to be configured with
different measurement ranges. The analyzer software calls these two ranges LOW and
HIGH. The LOW range setting corresponds with the analog output labeled A1 on the
rear panel of the instrument. The HIGH range setting corresponds with the A2 output.
While the software names these two ranges LOW and HIGH, they do not have to be
configured that way.
Also, in this mode the RANGE Test function displayed on the front panel during
SAMPLE mode will be replaced by two separate functions, RANGE1 & RANGE2.
LOW range = RANGE1 = Range value for output A1 = 0-1500 ppb
H
2
S.
HIGH range = RANGE2 = Range value for output A2 = 0-500 ppb
H
2
S.
For T101’s configured to measure both SO
2
and H
2
S in multigas measurement mode:
LOW range = RANGE1 = Range value for output A1= 0-1500 ppm
SO
2
.
HIGH range = RANGE2 = Range value for output A2 =0-1000 ppm
H
2
S.
To select the independent reporting range mode and set the upper measurement limits for
the two outputs, press:
SAMPLE ENTER SETUP PASS : 818
8 1 8 ENTR EXIT
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SETUP X.X RANGE CONTROL MENU
MODE
SET UNIT EXIT
SETUP X.X RANGE MODE: SNGL
SNGL IND AUTO ENTR EXIT
SETUP X.X RANGE CONTROL MENU
MODE
SET UNIT EXIT
SETUP X.X HIGH RANGE: 500.0 Con
c
0 0 5 0 0 .0 ENTR EXIT
SETUP X.X LOW RANG
E
: 500.0 Conc
0 0 1 0 0 .0 ENTR EXIT
SETUP X.X RANGE CONTROL MENU
MODE
SET UNIT EXIT
EXIT Returns
to the Main
SAMPLE Display
SETUP X.X RANGE MODE: DUAL
SNGL
IND AUTO ENTR EXIT
Toggle the
buttons under
each digit to set
the upper limit
of each range.
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL SETUP
NOTE
In INDEPENDENT range mode the two reporting ranges have separate slopes and offsets
for computing concentration and MUST be independently calibrated.