Teledyne T101 TV Converter Box User Manual


 
Appendix C
Warranty/Repair Questionnaire
T101, M101E
(05494D DCN5798)
C-1
Company: _________________________ Contact Name: _____________________________
Phone Number: ___________ Fax Number: _____________ Email: ____________________
Site Address: __________________________________________________________________
Can we connect to the instrument? If so, provide IP address or modem #:___________________
Model Serial Number: _________________________ Firmware revision: _________________
The serial number can be found on the back of the instrument, the firmware revision is displayed in the upper left corner of the
display when pressing SETUP on the front panel (Example: C.3).
1. List all front panel error/warning messages:_________________________________________
______________________________________________________________________________
2. Please complete the following table: (Depending on options installed, not all test parameters
shown below may be available in your instrument)
PARAMETER
RECORDED
VALUE
ACCEPTABLE
VALUE
PARAMETER
RECORDED
VALUE
ACCEPTABLE
VALUE
RANGE
ppb/ppm
50 ppb - 20 ppm SO2 SLOPE
1.0 ± 0.3
H2S STB
ppb
1 ppb with zero air
SO2 OFFS
mV
< 250
SAMP FL
cm³/min
600 ± 75 H2S SLOPE
PRES
IN-HG-A
~5”<ambient H2S OFFS
mV
< 250
PMT signal
with zero air
mV
-20 to 150 HVPS
V
400-900
PMT signal at
span gas conc
mV
ppb/ppm
0-5000
0-20 000 ppb
RCELL TEMP
°C
50 ± 1
NORM PMT at
span gas conc
mV
ppb/ppm
0-5000
0-20 000 ppb
BOX TEMP
°C
Ambient + ~5
UV LAMP
mV
1000 to 4800 PMT TEMP
°C
7 ± 2
LAMP RATIO
%
30-120% IZS TEMP
°C
50 ± 3
STR. LGT
ppm
100 ppb/ zero air
CONV TEMP
°C
315 ± 5
DARK PMT
mV
-50 to 200 ETEST (DIAG menu)
mV
2000 ± 1000
DARK LAMP mV
-50 to 200 OTEST (DIAG menu)
mV
2000 ± 1000
3. Has the analyzer been checked for leaks? Yes No For proper flows? Yes No
4. What are the failure symptoms? _________________________________________________
______________________________________________________________________________
______________________________________________________ Continue on back if necessary
5. Which tests have you done trying to solve the problem? _______________________________
______________________________________________________________________________
________________________________________________________
Continue on back if necessary
6. If possible, fax a portion of a strip chart or email a data file to customer service.
CUSTOMER SERVICE CONTACT INFORMATION: 6565 Nancy Ridge Drive, San Diego, CA 92121.
PHONE
: +1 858 657 9800 or 1-800 324 5190. FAX: +1 858 657 9816. EMAIL: api-customerservice@teledyne.com.
You can access and submit an online version of this form at http://www.teledyne-api.com/forms/p-fm101e.asp
07266B DCN6845