Teledyne API – T101 Operation Manual Operating Instructions
91
V
+DC Gnd
V OUT +
V OUT -
V IN +
V IN -
Recording
Device
ANALYZER
See Table 3-1 for
pin assignments
of Analog Out
connector on the
rear panel
Figure 4-4. Setup for Calibrating Analog Outputs
To make these adjustments, the AOUT auto-calibration feature must be turned off
(Section6.9). Activate the ANALOG I/O CONFIGURATION MENU (see Section
4.6.3), then press:
Press SET> to select the analog output channel to be configured:
DISPLAYED AS = CHANNEL
CONC_OUT_1 = A1
CONC_OUT_2 = A2
TEST OUTPUT = A4
DIAG ANALOG I / O CONFIGURATION
PREV NEXT ENTR EXIT
DIAG AIO AOUTS CALIBRATED: NO
< SET SET> CAL EXIT
FROM ANALOG I/O CONFIGURATION MENU
DIAG AIO CONC_OUT_1 :5V, NO CAL
< SET SET> EDIT EXIT
DIAG AIO CONC_OUT_1 CALIBRATED: YES
< SET CAL EXIT
DIAG AIO CONC_OUT_1 RANGE: 5V
SET> EDIT EXIT
DIAG AIO CONC_OUT_2 CALIBRATED: NO
< SET CAL EXIT
DIAG AIO CONC_OUT_1 VOLT
–
Z : 0 mV
U100 UP10 UP DOWN DN10 D100 ENTR EXIT
Toggle to increase / decrease the analog output
by 100, 10 or 1 counts.
Continue adjustments until the voltage measured
at the output of the analyzer and/or the input of
the recording device matches the value in the
upper right hand corner of the display to the
tolerance specified.
The concentration display will not change. Only
the voltage reading of your voltmeter will change.
DIAG AIO CONC_OUT_1 VOLT
–
S : 4500 mV
U100 UP10 UP DOWN DN10 D100 ENTR EXIT
DIAG AIO CONC_OUT_1 REC OFS: 0 mV
< SET SET> EDIT EXIT
If AutoCal is ON, go to
Section 6.7.3
EXIT ignores the
new setting.
ENTR accepts the
new setting.
DIAG AIO CONC_OUT_1 AUTO CAL: OFF
< SET SET> EDIT EXIT
07266B DCN6485