4.4.5. SETUP – PASS: PASSWORD PROTECTION
The menu system provides password protection of the calibration and setup functions to
prevent unauthorized adjustments. When the passwords have been enabled in the PASS
menu item, the system will prompt the user for a password anytime a password-protected
function (e.g., SETUP) is selected. This allows normal operation of the instrument, but
requires the password (101) to access to the menus under SETUP. When PASSWORD is
disabled (SETUP>OFF), any operator can enter the Primary Setup (SETUP) and
Secondary Setup (SETUP>MORE) menus. Whether PASSWORD is enabled or disabled,
a password (default 818) is required to enter the VARS or DIAG menus in the
SETUP>MORE menu.
Table 4-6. Password Levels
PASSWORD LEVEL MENU ACCESS ALLOWED
Null (000) Operation All functions of the main menu (top level, or
Primary, menu)
101
Configuration/Maintenance
Access to Primary and Secondary SETUP
Menus when PASSWORD is enabled
818
Configuration/Maintenance
Access to Secondary SETUP Submenus VARS
and DIAG whether PASSWORD is enabled or
disabled.
To enable or disable passwords, press:
07266B DCN6485