Teledyne T101 TV Converter Box User Manual


 
Teledyne API – T101 Operation Manual Operating Instructions
59
Table 4-2. Test Functions Defined
DISPLAY PARAMETER UNITS DESCRIPTION
RANGE Range
- -
Range1
Range2
PPB, PPM,
UGM &
MGM
The full scale limit at which the reporting range of the analyzer’s
ANALOG OUTPUTS are currently set. THIS IS NOT the physical range
of the instrument.
If IND or AUTO Range modes have been selected, two RANGE
functions will appear, one for each range.
H2S STB
1
Stability PPB
Standard deviation of H
2
S Concentration readings. Data points are
recorded every ten seconds. The calculation uses the last 25 data
points.
SAMP FL Sample Flow
cm³/min
(cc/m)
The flow rate of the sample gas through the sample chamber.
PRES
Sample
Pressure
in-Hg-A
The current pressure of the sample gas as it exits the sample
chamber, measured after the sample chamber.
PMT PMT Signal mV
The raw output voltage of the PMT.
NORM
PMT
NORMALIZED
PMT Signal
mV
The output voltage of the PMT after normalization for
temperature/pressure compensation (if activated).
UV LAMP
Source UV
Lamp Intensity
mV
The output voltage of the UV reference detector.
LAMP
RATIO
UV Source
lamp ratio
%
The current output of the UV reference detector divided by the
reading stored in the CPU’s memory from the last time a UV Lamp
calibration was performed.
STR. LGT Stray Light ppb
The offset due to stray light recorded by the CPU during the last zero-
point calibration performed.
DRK PMT Dark PMT mV
The PMT output reading recorded the last time the UV source lamp
shutter was closed.
DRK LMP
Dark UV
Source Lamp
mV
The UV reference detector output reading recorded the last time the
UV source lamp shutter was closed.
SO2
SLOPE
1
SO
2
measurement
Slope
-
The sensitivity of the instrument as calculated during the last
calibration activity. The slope parameter is used to set the span
calibration point of the analyzer.
SO2
OFFS
1
SO
2
measurement
Offset
mV
The overall offset of the instrument as calculated during the last
calibration activity. The offset parameter is used to set the zero point
of the analyzer response.
H2S
SLOPE
1
H
2
S
measurement
Slope
-
The sensitivity of the instrument as calculated during the last
calibration activity. The slope parameter is used to set the span
calibration point of the analyzer.
H2S
OFFS
1
H
2
S
measurement
Offset
mV
The overall offset of the instrument as calculated during the last
calibration activity. The offset parameter is used to set the zero point
of the analyzer response.
HVPS - - V
The PMT high voltage power supply.
RCELL
TEMP
Sample
Chamber Temp
°C
The current temperature of the sample chamber.
BOX
TEMP
Box
Temperature
°C
The ambient temperature of the inside of the analyzer case.
PMT
TEMP
Pmt
Temperature
°C
The current temperature of the PMT.
IZS
TEMP
1
IZS
Temperature
1
°C
The current temperature of the internal zero/span option. Only
appears when IZS option is enabled
CONV
H
2
S SO
2
°C
The current temperature of the catalytic converter that changes the
07266B DCN6485