7.1.7. SUMMARY OF QUALITY ASSURANCE CHECKS
The following items should be checked on a regularly scheduled basis to assure high
quality data from the T101. See Table 7-3 for a summary of activities. Also the QA
Handbook should be checked for specific procedures.
Table 7-3. Activity Matrix
CHARACTERISTIC ACCEPTANCE LIMITS
FREQUENCY AND
METHOD OF
MEASUREMENT
ACTION IF REQUIREMENTS ARE
NOT MET
Shelter
Temperature
Mean temperature between
22
o
C and 28
o
C (72
o
and 82
o
F),
daily fluctuations not greater
than ±2
o
C
Check thermograph
chart weekly for
variations greater
than ±2
o
C (4
o
F)
Mark strip chart for the affected
time period
Repair or adjust temperature
control
Sample
Introduction
System
No moisture, foreign material,
leaks, obstructions; sample line
connected to manifold
Weekly visual
inspection
Clean, repair, or replace as
needed
Recorder Adequate ink & paper
Legible ink traces
Correct chart speed and range
Correct time
Weekly visual
inspection
Replenish ink and paper supply
Adjust time to agree with clock;
note on chart
Analyzer
Operational
Settings
TEST measurements at nominal
values
2. T101 in SAMPLE mode
Weekly visual
inspection
Adjust or repair as needed
Analyzer
Operational
Check
Zero and span within tolerance
limits as described in Section.
9.1.3 of Sec. 2.0.9 (Q.A.
Handbook)
Level 1 zero/span
every 2 weeks; Level
2 between Level 1
checks at frequency
desired analyzer by
user
Find source of error and repair
After corrective action, re-
calibrate analyzer
Precision Check Assess precision as described in
Sec. 2.0.8 and Section. 3.4.3
(Ibid.)
Every 2 weeks,
Section. 3.4.3 (Ibid.)
Calc, report precision, Sec. 2.0.8
(Ibid.)
7.2. LEVEL 1 CALIBRATIONS VERSUS LEVEL 2 CHECKS
Essential to quality assurance are scheduled checks for verifying the operational status of
the monitoring system. The operator should visit the site at least once each week. It is
recommended Level 1 zero and span check conducted on the analyzer every two weeks.
Level 2 zero and span checks should be conducted at a frequency desired by the user.
Definitions of these terms are given in Table 7-4.
In addition, an independent precision check between 0.08 and 0.10 ppm must be carried
out at least once every two weeks. Table 7-3 summarizes the quality assurance activities
for routine operations. A discussion of each activity appears in the following sections.
To provide for documentation and accountability of activities, a checklist should be
compiled and then filled out by the field operator as each activity is completed.