Teledyne API T/E-Series Models 101, 102, and 108 Software Documentation (PN05492D DCN6485)
A-36
4 PMT temperature warning
5 Reaction cell temperature warning
6 Sample pressure warning
7 HVPS warning
8 System reset warning
9 Rear board communication warning
10 Relay board communication warning
11 Front panel communication warning
12 Analog calibration warning
13 Dynamic zero warning
14 Dynamic span warning
15 Invalid concentration
16 In zero calibration mode
17 In span calibration mode
18 In multi-point calibration mode
19 System status is OK (same meaning as SYSTEM_OK I/O signal)
20 Sample flow warning
21 IZS temperature warning
22
1
Converter temperature warning
23
5
Oxygenator flow warning
24
2
SO
2
concentration alarm limit #1 exceeded
25
2
SO
2
concentration alarm limit #2 exceeded
26
2
H
2
S/TRS concentration alarm limit #1 exceeded
27
2
H
2
S/TRS concentration alarm limit #2 exceeded
28
2+4
TRS concentration alarm limit #1 exceeded
29
2+4
TRS concentration alarm limit #2 exceeded
MODBUS Coil Registers
(single-bit; read/write)
0 Maps to relay output signal 36 (MB_RELAY_36 in signal I/O list)
1 Maps to relay output signal 37 (MB_RELAY_37 in signal I/O list)
2 Maps to relay output signal 38 (MB_RELAY_38 in signal I/O list)
3 Maps to relay output signal 39 (MB_RELAY_39 in signal I/O list)
20
3
Triggers zero calibration of range #1 (on enters cal.; off exits cal.)
21
3
Triggers span calibration of range #1 (on enters cal.; off exits cal.)
22
3
Triggers zero calibration of range #2 (on enters cal.; off exits cal.)
23
3
Triggers span calibration of range #2 (on enters cal.; off exits cal.)
07266B DCN6485