Teledyne T101 TV Converter Box User Manual


 
Teledyne API T/E-Series Models 101, 102, and 108 Software Documentation (PN05492D DCN6485)
A-18
Setup Variable
Numeric
Units
Default
Value
Value Range Description
REMOTE_CAL_MODE — SO2-LOW SO2-LOW,
SO2-HIGH,
H2S-LOW,
H2S-HIGH,
TRS-LOW
6
,
TRS-HIGH
6
Gas and range to calibrate
during contact-closure and
Hessen calibration. Enclose
value in double quotes (“)
when setting from the RS-
232 interface.
HOLD_DAC_ON_CAL OFF OFF, ON ON holds D/A outputs during
zero/span calibration; OFF
permits D/A outputs to
change.
PASS_ENABLE OFF OFF, ON ON enables passwords; OFF
disables them.
RCELL_CYCLE Seconds 2 0.5–30 Reaction cell temperature
control cycle period.
RCELL_PROP 1/ºC 0.3 (prop.
band = 3.3
ºC)
0–10 Reaction cell temperature
PID proportional coefficient.
RCELL_INTEG — 0.005 0–10 Reaction cell temperature
PID integral coefficient.
RCELL_DERIV — 0.5 0–10 Reaction cell temperature
PID derivative coefficient.
IZS_CYCLE Seconds 2 0.5–30 IZS temperature control
cycle period.
IZS_PROP 1/ºC 1 (prop.
band = 1 ºC)
0–10 IZS temperature PID
proportional coefficient.
IZS_INTEG 0.03 0–10 IZS temperature PID integral
coefficient.
IZS_DERIV 0 0–10 IZS temperature PID
derivative coefficient.
650 HVPS_SET Volts
Warnings:
400–900
0–2000 High voltage power supply
warning limits. Set point is
not used.
MAX_PMT_DETECTOR mV 4995 0–5000 PMT detector maximum
warning limit.
450 PHOTO_ABS_LIMITS mV
Warnings:
125–625
0–5000 Pre-amplified UV lamp
minimum/maximum warning
limits. Set point is not used.
3500 UV_LAMP_LIMITS mV
Warnings:
1000–4995
0–5000 UV lamp minimum/maximum
warning limits. Set point is
not used.
ELEC_TEST_LEVEL
9
0 0–65535 Electrical test level setting.
OPTIC_TEST_LEVEL
9
0 0–65535 Optical test level setting.
07266B DCN6485