Teledyne API – T101 Operation Manual Operating Instructions
63
The default status of the calibration password is OFF. To enable the calibration password
press:
SAMPLE ENTER SETUP PASS : 818
8 1 8
ENTR
EXIT
SAMPLE RANGE = 500.0 PPB H2S =X.XXX
< TST TST > CAL SETUP
SETUP X.X
PRIMARY SETUP MENU
CFG DAS RNGE
PASS
CLK MORE EXIT
SETUP X.X
CAL. PASSWORD ENABLE: OFF
OFF ENTR EXIT
SETUP X.X PASSWORD ENABLE:
ON
ON ENTR EXIT
SETUP X.X PASSWORD ENABLE: ON
ON ENTR EXIT
ENTR
accepts
the change
EXIT
ignores
the change
Toggles
password
status On/Off
ENTR accepts
displayed
password value
EXIT returns to
SAMPLE
display
CAL. PASSWORD
default state is
OFF
If the calibration password (101) is enabled, the following menu button sequence will be
required to enter one of the calibration modes:
SAMPLE ENTER SETUP PASS : 0
0 0 0 ENTR EXIT
SAMPLE RANGE = 500.0 PPB H2S =X.XXX
< TST TST > CAL CALZ CALS SETUP
M-P CAL RANGE = 500.0 PPB H2S =X.XXX
< TST TST > ZERO CONC EXIT
Prompts
password
number
SAMPLE ENTER SETUP PASS : 0
1 0 1 ENTR EXIT
Press
individual
buttons to set
101
Continue calibration process …
07266B DCN6485