Calibration Procedures Teledyne API – T101 Operation Manual
176
STEP TWO: Perform the zero/span check.
Wait until H2S
STB falls below
0.5 ppb. This may
take several
minutes.
The value of H2S
STB may jump
significantly. Wait
until H2S STB falls
below 0.5 ppb. This
may take several
minutes.
SPAN CAL M H2S STB=XXX.X PPB H2S =XXX.X
< TST TST > ZERO SPAN CONC EXIT
SAMPLE H2S STB=XXX.X PPB H2S =XXX.X
< TST TST > CAL CALZ CALS SETUP
ZERO CAL M H2S STB=XXX.X PPB H2S =XXX.X
< TST TST > ZERO CONC EXIT
SAMPLE H2S STB=XXX.X PPB H2S =XXX.X
< TST TST > CAL CALZ CALS SETUP
EXIT returns to the mai n
SAMPLE dis
p
la
y
ACTION:
Record the
H
2
S readings
presented in the
upper right corner of
the display.
ACTION:
Record the
H
2
S readings
presented in the
upper right corner of
the display.
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL CALZ CALS SETUP
Scroll to the H2S
STB test function.
07266B DCN6485