Teledyne T101 TV Converter Box User Manual


 
Teledyne API – T101 Operation Manual Calibration Procedures
181
The following example sets Sequence2 to carry out a zero-span calibration every other
day starting at 01:00 on September 4, 2002, lasting 15 minutes. This sequence will start
0.5 hours later each day.
Table 6-4. Example Auto-Cal Sequence
MODE AND
ATTRIBUTE
VALUE COMMENT
SEQUENCE 2 Define Sequence #2
MODE ZERO-SPAN Select Zero and Span Mode
TIMER ENABLE ON Enable the timer
STARTING DATE Sept. 4, 2002 Start after Sept 4, 2002
STARTING TIME 01:00 First Span starts at 01:00
DELTA DAYS 2 Do Sequence #2 every other day
DELTA TIME 00:30 Do Sequence #2 0.5 h later each day
DURATION 15.0 Operate Span valve for 15 min
CALIBRATE ON The instrument will re-set the slope and offset values for the H
2
S
channel at the end of the AutoCal sequence
07266B DCN6485