Operating Instructions Teledyne API – T101 Operation Manual
84
4.6.2. ANALOG OUTPUT STEP TEST
This test can be used to check the accuracy and proper operation of the analog outputs.
The test forces all four analog output channels to produce signals ranging from 0% to
100% of the full scale range in 20% increments. This test is useful to verify the operation
of the data logging/recording devices attached to the analyzer.
To begin the Anog Output Step Test press:
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL SETUP
SAMPLE ENTER SETUP PASS : 818
8 1 8 ENTR EXIT
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
DIAG SIGNAL I / O
NEXT ENTR EXIT
Exit-Exit
returns to the
DIAG menu
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
DIAG
A
NALOG OUTPUT
PREV NEXT ENTR EXIT
DIAG AOUT ANALOG OUTPUT
0% EXIT
Performs
analog output
step test.
0% - 100%
DIAG AOUT ANALOG OUTPUT
[0%] EXIT
Pressing the button under “0%” while performing the test will pause
the test at that level. Brackets will appear around the value; example:
[20%] Pressing the same button again will resume the test.
07266B DCN6485